We have a problem with the agfa CR 35 X . it gives the error code 20442 , and the image looks distorted with a lot of horizontal lines .
This Error occurs occasionally .

Log inor Register

to rate and post a comment

Zhang r.s

re: Please help with Agfa CR35 X

March 13, 2018 09:06

(1) Enter the Service menu

(2) Select
4 Show error
1 Explain error code

(3) Select in error code history corresponding entry
(4) Reason, meaning and cure of error are listed

Log inor Register

to rate and post a comment

(319)
(59)
(2)

juan pablo amado

re: Please help with Agfa CR35 X

March 13, 2018 09:07

20442
E_SCD_SCB_DMA_BUF_OVERFLOW
Meaning : There was at least one DMA ring buffer overflow during scan.
Reason : The image data was not stored away quick enough by the
following buffer acquisition (BAC) software.
Bad disk, system overload, ...

Log inor Register

to rate and post a comment

(13)

hasan raza

re: Please help with Agfa CR35 X

May 30, 2018 09:40

contact me my email address faisal.medical.equipments@gmail.com

Log inor Register

to rate and post a comment

THE (LEADER) IN MEDICAL IMAGING TECHNOLOGY SINCE 1982. SALES-SERVICE-REPAIR

Special-Pricing Available on Medical Displays, Patient Monitors, Recorders, Printers, Media, Ultrasound Machines, and Cameras. This includes Top Brands such as SONY, BARCO, NDS, NEC, LG, EDAN, EIZO, ELO, FSN, PANASONIC, MITSUBISHI, OLYMPUS, & WIDE.

 

advertisement

 

advertisement
advertisement
El acceso y el uso de este sitio está conforme a los términos y a las condiciones de nuestro AVISO LEGAL & AVISO DE LA AISLAMIENTO
Característica de y propietario DOTmeda .com, inc. Copyright ©2001-2018 DOTmed.com, Inc.
TODOS LOS DERECHOS RESERVADOS