SEARCH
Ubicación actual:
>
>
> GE High Speed CT ( urgent)

GE High Speed CT ( urgent)

Posted by an Anonymous User

GE High Speed CT ( urgent)

December 06, 2017 08:52

I’m Getting Gantry Error .Code: 13-1002-09 error detected by:TGP(go) gantry cover switch is open and axial motor is not Powered .Please give me solution, it’s very urgent

Log inor Register

to rate and post a comment

(9)
(4)
(5)

Bruce Hull

re: GE High Speed CT ( urgent)

December 06, 2017 09:24

There are switches for both the front and back covers. They are located on the right side of the gantry. There is also two switches for the right side cover. They are all in series.

Log inor Register

to rate and post a comment

(120)
(8)
(5)

Manuel Esquer

re: GE High Speed CT ( urgent)

December 06, 2017 09:24

CHECK MICROSWITCH RIGHT SIDE COVER

Log inor Register

to rate and post a comment

(125)
(12)
(2)

Lin Dong

re: GE High Speed CT ( urgent)

December 06, 2017 09:27

Please check the Gantry cover which is not properly installed or you need to by-pass the switch when you do the service and test

Log inor Register

to rate and post a comment

(135)
(16)

CRAIG WAMBURA

re: GE High Speed CT ( urgent)

December 07, 2017 09:25

1. Check alignment between Gantry cover and Gantry Cover Switch;
2. Check gantry Cover Switch and its cabling to TGP

Log inor Register

to rate and post a comment

Laser-Tech.com: Used Lasers & IPL - Buy, Sales, Repair, Parts 855-LASER-00

Are you looking to Buy or Sell a Laser? Do you need Service, Repair or Parts for you Laser or IPL system? Call LASER-TECH today at 855-527-3700 We Sell and Service all major brands and manufactures!

 

 

El acceso y el uso de este sitio está conforme a los términos y a las condiciones de nuestro AVISO LEGAL & AVISO DE LA AISLAMIENTO
Característica de y propietario DOTmeda .com, inc. Copyright ©2001-2017 DOTmed.com, Inc.
TODOS LOS DERECHOS RESERVADOS