February 03, 2020 09:56

BV Endura Error M032 Can you help me

Log inor Register

to rate and post a comment

(125)
(18)
(15)

Doscar Ernesto Perez Pereyra

re: BV Endura

February 04, 2020 10:27

Error M032 are parameters reset. Need Bv Scope for take out the error with the calibrations.

Log inor Register

to rate and post a comment

(22)
(6)

Sergey Radionov

re: BV Endura

February 04, 2020 10:29

032. Adjustment status: not verified.
Logfile text: X-ray tube performance failed or never performed. Perform the following actions via BV Scope: Install
- adjustment - X-ray tube - Filament. Install - Performance Test - X-ray tube performance.

Log inor Register

to rate and post a comment

(47)
(4)

Willi

re: re: BV Endura

February 05, 2020 10:18

Dear all,

Yes this is correct, but mostly when you have this error then the system has more adjustment problem.
Then the Standcontroller will not only have no Tube data lost.

Log inor Register

to rate and post a comment

(62)
(17)

tim cowell

re: BV Endura

February 04, 2020 10:30

You need to complete a successful performance test.