> Agfa Drystar 3000 Images too dark

Agfa Drystar 3000 Images too dark

(57)
(2)

Svilen Popov

Agfa Drystar 3000 Images too dark

October 11, 2017 01:00

Hello!
I have problem with printer Agfa Drystar 3000. It is connected via Dicom with MR, and everything works, but the images are dark. When I perform FILM Sensito calibration.... it passes and completes, but have warning:"MDM-Sens not Okay" and doesn't apply it(the last calibration witch is applied is 7 years ago). I perform and MDM calibration - it's pass too, completed without errors. Try again to perform Film Sensito calibration...the same result("MDM-Sens not Okay").
And I don't know how to light the images...

Log inor Register

to rate and post a comment

(53)
(8)
(2)

Gery Rodriguez

re: Agfa Drystar 3000 Images too dark

October 12, 2017 01:16

Sounds like you have a temperature problem. Therefore affecting contrast. Check your heaters

Log inor Register

to rate and post a comment

(7)
(1)

Michele Di Lello

re: Agfa Drystar 3000 Images too dark

October 12, 2017 01:18

check the expiration date on the film

Log inor Register

to rate and post a comment

(24)
(7)

Vo Quoc

re: Agfa Drystar 3000 Images too dark

October 12, 2017 01:18

Hi Friend,
You ảre doing well
Its not problem with printer ds3000
Its cause from the film
You are using wrong film type for ds3000
Correct film type for ds3000 is DT1B but its discontinue.
So currently you are loading DT2B for this printer ds3000 right ?
DT2B loading to ds3000 , printer can print but images more darkend than DT1B

Thank you
Bao

1 person liked this
Log inor Register

to rate and post a comment

(14)
(4)

RODRIGO CORTeS

re: Agfa Drystar 3000 Images too dark

October 12, 2017 01:18

What kind of film are you printing... If it's DT2, that the reason

Log inor Register

to rate and post a comment

(57)
(2)

Svilen Popov

re: re: Agfa Drystar 3000 Images too dark

October 12, 2017 09:53

Hm... yes I use DT2B - obviously this is the reason.
Thank you everyone for answers!!!

P.S. Is there any settings from IMOS to light a little bit images ??? I tried with Window/Level at Keypad 4(local keypad), but I am not sure, if it change something....

Log inor Register

to rate and post a comment

(57)
(2)

Svilen Popov

re: re: re: Agfa Drystar 3000 Images too dark

October 13, 2017 09:29

I am not sure if the films are DT2B or DT5000B, but it is not DT1B

Log inor Register

to rate and post a comment

(112)
(12)

CRAIG WAMBURA

re: Agfa Drystar 3000 Images too dark

October 16, 2017 10:41

1. Could be dirt on the infrared sensor. Through the dirt the temperature sensor measures a low temperature which then results in a higher dercurl temperature.
Contamination due to light particles which are shed by the film; in which case white dust may be found in the area of the decurl unit and the thermal head.

2. IR sensor could be defective

3. PMC1 Board could be defective

Log inor Register

to rate and post a comment

ADMAR NEURODIAGNOSTIC, INC: INSTRUMENT SALES & SERVICE 1-800-331-3957

We sell/service EMG-EEG-IOM epuipment with in-house trained technicians. With over 30 years experience we offer sales/service to Xltek/Natus Neuromax, Cadwell systems,Teca Synergy,Sapphire,Nicolet Quest EMG,Dantec EMG.1-800-331-3957/sales@admarneuro.com DOTmed Certified

 

 

El acceso y el uso de este sitio está conforme a los términos y a las condiciones de nuestro AVISO LEGAL & AVISO DE LA AISLAMIENTO
Característica de y propietario DOTmeda .com, inc. Copyright ©2001-2017 DOTmed.com, Inc.
TODOS LOS DERECHOS RESERVADOS