Hi
Go to offline scan . Do scan without xray . Plot it or display the data . Try to locate the channel zero. Swap the board and repeat procedure. If the channel moves than change the board if not change the detector.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
(18)
(2)
Pepe Garcia
X–ray Zero Detect GE CT SCAN PROSPEED
July 01, 2015 10:28
GOOD DAY I HAVE THE PROBLEM OF X-ray Zero Detect PROSPEED GE IN EQUIPMENT AND CHECK THE BOARDS anybody CAM have some coment?
to rate and post a comment
yao wen hsu
re: X–ray Zero Detect GE CT SCAN PROSPEED
July 01, 2015 11:30
Xenon gas type ? If is this, may be check the 1000Vdc power supply
to rate and post a comment
(6)
(2)
Charbel Tadros
re:X–ray Zero Detect GE CT SCAN PROSPEED
July 02, 2015 05:22
Try to know wich channel is down to zero , then sawp boards. If is fix then you have a detector problem
to rate and post a comment
(6)
(2)
Charbel Tadros
re: X–ray Zero Detect GE CT SCAN PROSPEED
July 04, 2015 08:21
Hi
Go to offline scan . Do scan without xray . Plot it or display the data . Try to locate the channel zero. Swap the board and repeat procedure. If the channel moves than change the board if not change the detector.
to rate and post a comment