roche integra 400 problem

(4)
(1)

samer ali

roche integra 400 problem

September 29, 2019 04:30

hi.
i need help with integra 400 problem.
after turning on the device .it initialize probes and wash them successfully.after that the arm goes toward prob initializing post (which used in probes calibration)moves a little down in z direction the this message appears on pc monitor:
a hardware problem has been detected in the z-transfer controller
message ID : 14.3232.02.068.086
the arm moves in z direction successfully during initializing and washing steps which means the problem isn't in the z controller or motors
i tried to replace z-controller but the problem is still there.can some one help me
thanks

Log inor Register

to rate and post a comment

(1)
(1)

Maurizio Fasolino

re: roche integra 400 problem

October 01, 2019 09:43



Hi,

message ID : 14.3232.02.068.086

Involved Controller Board:
02=Transfer Z Control

Error Description:
068=InitPostNotFound
During initialization a probe is moved against an Init Posts using the Level Detection circuitry. The interrupt signal "Level found" did not occur. The Init Post could not be detected correctly.

Command Description:
086=PositionMotorZCLevDet
Moves the Probe C in Z direction using the Level Detection circuitry.The motor stops when level is found. The command parameters include the allowed Zmax movement. The position, where niveau is found, is read out.

Possible Causes:
02.068.086
• ZC-Motor mechanical blockage
• Tip of probe C is damaged
• ISE Silicon Tube leading from Mixing Tower to Valve V74 is not correctly installed. Tube should be installed in the tube guide behind the ISE Mixing Tower.
PCB ABS Photometer (Slave 03)

Good day

1 person liked this
Log inor Register

to rate and post a comment

(4)
(1)

samer ali

re: re: roche integra 400 problem

October 04, 2019 10:03

thanks Maurizio Fasolino
your reply was very helpful.the problem is solved.

Log inor Register

to rate and post a comment