Radiometer ABL 800 hemolyzer failure

Tags: Clinical Lab, RADIOMETER

(5)
(2)

Ciprian Lupu

Radiometer ABL 800 hemolyzer failure

November 14, 2014 08:31

Hello everyone!
I got a problem with my Radiometer ABL800. The Hemolyzer test in Service menu shows these results:
1. Thermostat temperature: Passed; 37.2 degrees
2. Thermostat heat: Passed 12.5%
3. Tuning voltage: Failed 67.0 Vpeak
4. Tuning Frequency: Passed 30.284khz
5. Impedance shape: Passed
6. Cavitation absorbance: Failed Max 4mAbs
7. Cavitation Drive Current Failed 100%

I believe that "pass" tests are mainly related to the logicboard/controller and the "fail" tests are related to the mechanics of the hemolyzer.

What do you think? Is it possible to fix it?

Thank you :)

Log inor Register

to rate and post a comment

(165)
(37)
(3)

Dmitry Dederer

re: Radiometer ABL 800 hemolyzer failure

November 16, 2014 11:19

Hello! Do you have service manuals for this system? You need to check caviation. What kind of gas do you use?

Log inor Register

to rate and post a comment

(5)
(2)

Ciprian Lupu

re: re: Radiometer ABL 800 hemolyzer failure

November 20, 2014 06:30

Hello,I do have a service manual for this system but it does not contain any info about the hemolyzer. Maybe Radiometer keeps this kind of info only for themselves :-) Unfortunately I cannot understand your question . Are you refering to the calibration gas? Have a nice day

Log inor Register

to rate and post a comment

Lenny Place

re: re: Radiometer ABL 800 hemolyzer failure

July 22, 2015 11:33

Dmitry
do you have the service manual for this unit? i am trying to help someone and the manual would be helpful
thank you