Hello All,
i have this machine gives error 61 in the stand by mode after power on,
if any one can help me with manuals and how can i fix this error,
thanks
with best regards

1 person liked this
Log inor Register

to rate and post a comment

(57)
(1)

Abd Ullah

re: Planmeca proline PM2002

March 14, 2019 04:26

Dear mohamed
This error code for processor communication
I can help you
abdlaamin0@gmail.com

Log inor Register

to rate and post a comment

(28)
(3)

Matt Stiern

re: Planmeca proline PM2002

March 14, 2019 04:43

ERROR61:Generator Processor does not respond:
+33V below 24v
serial communication down
no power to generator processor bd
+5v shorted
Faulty SW Prom
Board defective
check cables and voltages disconnect motors and sensors