Phillip Intellvue MP 50

July 18, 2019 09:13

I need help with solving the ERROR MESSAGE : NIBP CUFF OCCLUSION

Log inor Register

to rate and post a comment

(2)

Greg Varcas

re: Phillip Intellvue MP 50

July 19, 2019 10:43

first thing is to swap out the cuff and hose to see if the problem is there. if not, an internal crimped hose. if these two things don't help possibly the pressure sensor on the board is defective.

Log inor Register

to rate and post a comment

(36)
(7)

Ray Brown

re: Phillip Intellvue MP 50

July 19, 2019 10:45

There is no specific error that says exactly what you said. The word Occlusion does not appear as an Alarm or INOP. The only thing that comes close is if you go into the Setup of the NBP and select the VeniPuncture tab, which will inflate the cuff up to the last diastolic pressure and hold it for 3 minutes, to assist taking blood samples from that arm.
Please restate the entire message, especially if it has a ! or * or ** in the text.

Log inor Register

to rate and post a comment

(79)
(11)

Steve Orcel

re: Phillip Intellvue MP 50

July 19, 2019 10:46

Hi
Did you change the Pump ?

Log inor Register

to rate and post a comment

(36)
(7)

Ray Brown

re: re: Phillip Intellvue MP 50

July 21, 2019 12:57

That would be if it said "NBP Equip Malf"..

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Pioneer Biomedical – The #1 Solution for All Your Patient Monitor Needs!

Our expert team provides prompt, dependable solutions for rentals, sales and repairs on patient monitors, gas modules, telemetry and more! We focus on producing superior results and providing first-class customer service. Call (888) 644-3274.

stats
advertisement

Próximos Eventos de Subasta

NorthWest Supply
TIMED
Apr 08, 2024 (99 lots)