The Point will be, that the Gantry has a BIST Error and the he don't start the communication with the AWS.
Have you look to the Gantry Webserver and his Logfile ?
On the left side is the stand controller and there are a 7 dot
display with the startup count. May there you can see on
whats startup position he stop.
The same display is by the arm behind the compression Unit.
Then you can look by the service manual what this can be.
Or, you ask a Siemens ing. for the manual. Then the gantry
is the same as the Siemens one.
The Emergency stop must be correct. You can also with
jumpers on the stand controller switch of the function of it.
(2)
MohamedElfandi
Philips MammoDiagnost DR 2.0
May 05, 2020 10:06
Any body can help me about philips MammoDiagnost DR 2.0
to rate and post a comment
(13)
Herrmann
re: Philips MammoDiagnost DR 2.0
May 06, 2020 09:45
Hi Mohamed,
When you not write, what your problem with the Philips
MammoDr is, then we all can't help you ..
Kind regards
Willi
to rate and post a comment
(2)
MohamedElfandi
re: re: Philips MammoDiagnost DR 2.0
May 07, 2020 11:00
The problem not have communication with host gantry state is Halt .
Self test collimation done
But grid not movement in self test
And Emergency stop is active
Grid not movement in self test
I would like how i can change gantry to initial mode
to rate and post a comment
(13)
Herrmann
re: re: re: Philips MammoDiagnost DR 2.0
May 08, 2020 11:00
Hi Mohamed,
The Point will be, that the Gantry has a BIST Error and the he don't start the communication with the AWS.
Have you look to the Gantry Webserver and his Logfile ?
On the left side is the stand controller and there are a 7 dot
display with the startup count. May there you can see on
whats startup position he stop.
The same display is by the arm behind the compression Unit.
Then you can look by the service manual what this can be.
Or, you ask a Siemens ing. for the manual. Then the gantry
is the same as the Siemens one.
The Emergency stop must be correct. You can also with
jumpers on the stand controller switch of the function of it.
Kind regards
Willi
to rate and post a comment
(2)
MohamedElfandi
re: re: re: re: Philips MammoDiagnost DR 2.0
June 29, 2020 12:55
in logfile gantry webserver given E_ERROR_COLLISION_DETECTION _ROT
number 109-1-0-0
to rate and post a comment