Dear Sirs need immediate assistance on a Philips Libra.
M050 Initialization error.
I have service manual, but it gives very little information concerning the error code.
Any help or direction would be greatly appreciated.
Thanks!!
Todd Starnes
828-320-0064
to rate and post a comment
Lenny Place
re: Philips BV Libra
March 19, 2015 01:41
the M050 is pretty generic check the SUCO and see if any of the boards has a red LED lite.
I can send you some more information if you give me an email address
to rate and post a comment
(4)
(28)
Moshe Alkalay
re: re: Philips BV Libra
May 07, 2020 10:59
err= M050 and in details M431 X ray generator not availaBLE.
Can u help?
to rate and post a comment
(2)
Steven Zhao
re: Philips BV Libra
March 21, 2015 04:50
M050 Means the communication problem between Stand and trolley, if the trolley startup ok ,mostly,some wires broken inside stand trolley cable or the serial number lost in system software configuration. if the monitor keep darkness ,the problem with DFI ,it should be replaced, expected, This can help you.
to rate and post a comment
(1)
muhammad ziad
re: Philips BV Libra
October 11, 2017 12:57
Dear Todd Starnes , can you please send me this manual as i need it deadly. ziyad2222pk@gmail.com