Hello everyone.
I have c-arm oec 7700 with problems.
when I turn on not seeing any mistake but does not appear any circle in Minitors.
When I press the switch of the fluoroscopy,the kv go to maximum value and the mA remains at zero.
some technical opinion?. We also need the service manual.
best regards.
thanks
Hello
You have no xray from the tube.
I suppose tsth you must check the xray tube and filament circuit.
Is a good idea to put betwin the II and xray tube an old radiographic cassete to see xray.
regards
uto
I want to check the circuit filament, but I do not have the service manual.
and also check that there is no radiation with the screen.
I need the service manual
thank you for your comment.
(17)
alejandro mendoza
OEC 7700 problem
August 06, 2008 07:57
Hello everyone.
I have c-arm oec 7700 with problems.
when I turn on not seeing any mistake but does not appear any circle in Minitors.
When I press the switch of the fluoroscopy,the kv go to maximum value and the mA remains at zero.
some technical opinion?. We also need the service manual.
best regards.
thanks
to rate and post a comment
Umberto Sciunnach
Re: OEC 7700 problem
August 08, 2008 02:01
Hello
You have no xray from the tube.
I suppose tsth you must check the xray tube and filament circuit.
Is a good idea to put betwin the II and xray tube an old radiographic cassete to see xray.
regards
uto
to rate and post a comment
(17)
alejandro mendoza
Re: OEC 7700 problem
August 12, 2008 11:48
I want to check the circuit filament, but I do not have the service manual.
and also check that there is no radiation with the screen.
I need the service manual
thank you for your comment.
to rate and post a comment
ali wesam
re: OEC 7700 problem
December 20, 2014 09:27
could send me the
to rate and post a comment
Mohammed Abdullah
re: OEC 7700 problem
December 26, 2014 11:55
i have the manuals if you need it contact me
magmesme@gmail.com
regards
to rate and post a comment