Hello, I have problems with a GE Senographe 800T, showing the following error: E12 Checksum Error 071/081. I changed the ram-back up battery, but had this error again. Also saw, that led "FPGA OK" on GENERATOR & ARM INTERFACE BOARD - flashes constantly , and haven't movements of the device. What I must to do ??? Any ideas ???
Reset the checksum and proceed to calibrate generator and arm. Can also load databases from another unit to verify everything works (before calibrating for 4~5 hours). Enjoy.
If You do not have the backup floppy of the system, You need recalibrate all (generator and arm), You need software of service, system manual, cable of conection, lap-top with serial port, and other accesories. and experience in the equipment.
I already begun calibrations and made generators calibration. If you could send me backup file with Arm calibrations, because I haven't this "force detector" to measure compression in daNs, and it will be very difficult without it(to calibrate the Arm)
I calibrated generator arm and aec, and now the device works. Now I must connected it to laptop and save backup file. I have one Mammocom for DMR, but didn't test it yet. Is it the same and for 800T ? And could someone send me Advanced Service Manual for Senographe 800T?
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
(15)
(1)
Svilen Popov
Mammo GE Senographe 800T E12 071/081
August 07, 2019 11:20
Hello, I have problems with a GE Senographe 800T, showing the following error: E12 Checksum Error 071/081. I changed the ram-back up battery, but had this error again. Also saw, that led "FPGA OK" on GENERATOR & ARM INTERFACE BOARD - flashes constantly , and haven't movements of the device. What I must to do ??? Any ideas ???
P.S. I haven't backup floppies of the system.
to rate and post a comment
Lucio Moraes
re: Mammo GE Senographe 800T E12 071/081
August 08, 2019 09:19
E12 Checksum Error Generator : Checksum error on Battery BackāUp RAM (BBR) 071/081
to rate and post a comment
(1)
Rich_SRS
re: Mammo GE Senographe 800T E12 071/081
August 08, 2019 09:20
Reset the checksum and proceed to calibrate generator and arm. Can also load databases from another unit to verify everything works (before calibrating for 4~5 hours). Enjoy.
to rate and post a comment
(2)
Raul Cervantes
re: Mammo GE Senographe 800T E12 071/081
August 08, 2019 09:21
If You do not have the backup floppy of the system, You need recalibrate all (generator and arm), You need software of service, system manual, cable of conection, lap-top with serial port, and other accesories. and experience in the equipment.
to rate and post a comment
(15)
(1)
Svilen Popov
re: Mammo GE Senographe 800T E12 071/081
August 08, 2019 09:25
Someone to have back-up copy with calibrated settings and parameters for this unit ?
to rate and post a comment
(2)
Raul Cervantes
re: re: Mammo GE Senographe 800T E12 071/081
August 09, 2019 10:01
Is possible but the automatic exposition will not work only the manual. I can send you the backup if you want
to rate and post a comment
(15)
(1)
Svilen Popov
re: re: re: Mammo GE Senographe 800T E12 071/081
August 11, 2019 02:23
I already begun calibrations and made generators calibration. If you could send me backup file with Arm calibrations, because I haven't this "force detector" to measure compression in daNs, and it will be very difficult without it(to calibrate the Arm)
to rate and post a comment
(15)
(1)
Svilen Popov
re: Mammo GE Senographe 800T E12 071/081
August 20, 2019 10:31
I calibrated generator arm and aec, and now the device works. Now I must connected it to laptop and save backup file. I have one Mammocom for DMR, but didn't test it yet. Is it the same and for 800T ? And could someone send me Advanced Service Manual for Senographe 800T?
to rate and post a comment