Innova 2000 booting problem.
Hi all, in doing the installation of a Innova 2000 with a dl6, and have problems with the booting sequence. On the Advantx Console show 4f2d then 4f2E and 4f48 (boot Failure), the windows nt Computer works ok (DL), the Titan Boot ok, and the Generator works ok, positioner boots ok, also on the monitor, the Position information Table, C-arm movements Shows â%+4.ofâ or similar on all data (when i connect with the Positioner service tool I can see all the right info)
Im thinking on Rtac or some communication problem.
Can anyone help me with this problem.
Hi 4f2d is problem COMUNICATION between RTAC and DETECTOR check power supply detector- check liquid in the CHILLER DETECTOR or is possibile DISP BOARD IN THE RTAC check led power SUPPLY at power on of led are ON and quicly off problem with some subsystem
Thanks for the info francesco, the detector power supply goes on and on the startup goes off, we already change the detector power supply. Wich subsystem we have to check??
Had the same detector communication problem on an install. My problem was when the detector was deinstalled, cooling water ran into the detector. I took the cover of the detector and found water. I rinsed it with alcohol (you can't break broken) and dried it with a hair dryer. It worked.
It is normally that power supply goes on then off during the start up then when it boots it turned on again.
Could you please send to me the error log of DL as well as the advantx on my mail:
Eng_mina_mounir@yahoo.com
Whrn power supply go on and after go off power supply is always good go off becsuse is in security normali reason is liquid in the CHILLER optical fiber or problem in the RTAC try test COMUNICATION rtac-detector
Hello,
Why you said it's booting problem??
It may be something else specially that error 4F2D does not relate to booting sequence.
Did you check the error log of DL?
I can help you to solve your issues
Hi we found broken fiber insite de detector. Now no 4f2d or 4f2e, but our 4f48 boot error its on advantx console. On tbe service home page i cant see the positioner release, Unknow is showing
SPECT, PET, CT and PET/CT equipment sales and service from OEM certified FSEs. Maintenance, repair, installation, relocation, and parts available with 24/7 support. 1-866-682-6331 or email sales@TTGimagingsolutions.com.
(8)
RobertCow
Innova 2000 booting problem
November 05, 2018 06:44
Innova 2000 booting problem.
Hi all, in doing the installation of a Innova 2000 with a dl6, and have problems with the booting sequence. On the Advantx Console show 4f2d then 4f2E and 4f48 (boot Failure), the windows nt Computer works ok (DL), the Titan Boot ok, and the Generator works ok, positioner boots ok, also on the monitor, the Position information Table, C-arm movements Shows â%+4.ofâ or similar on all data (when i connect with the Positioner service tool I can see all the right info)
Im thinking on Rtac or some communication problem.
Can anyone help me with this problem.
to rate and post a comment
(9)
Francesco Didonato
re: Innova 2000 booting problem
November 07, 2018 08:18
Hi 4f2d is problem COMUNICATION between RTAC and DETECTOR check power supply detector- check liquid in the CHILLER DETECTOR or is possibile DISP BOARD IN THE RTAC check led power SUPPLY at power on of led are ON and quicly off problem with some subsystem
to rate and post a comment
(4)
Roberto Angel Vaca Valverde
re: re: Innova 2000 booting problem
November 08, 2018 09:55
Thanks for the info francesco, the detector power supply goes on and on the startup goes off, we already change the detector power supply. Wich subsystem we have to check??
to rate and post a comment
(9)
Bruce Barton
re: re: re: Innova 2000 booting problem
November 10, 2018 11:09
Had the same detector communication problem on an install. My problem was when the detector was deinstalled, cooling water ran into the detector. I took the cover of the detector and found water. I rinsed it with alcohol (you can't break broken) and dried it with a hair dryer. It worked.
to rate and post a comment
(9)
Mina halim
re: re: re: Innova 2000 booting problem
November 11, 2018 11:15
It is normally that power supply goes on then off during the start up then when it boots it turned on again.
Could you please send to me the error log of DL as well as the advantx on my mail:
Eng_mina_mounir@yahoo.com
to rate and post a comment
(9)
Francesco Didonato
re: re: re: Innova 2000 booting problem
November 11, 2018 11:16
Whrn power supply go on and after go off power supply is always good go off becsuse is in security normali reason is liquid in the CHILLER optical fiber or problem in the RTAC try test COMUNICATION rtac-detector
to rate and post a comment
(9)
Mina halim
re: Innova 2000 booting problem
November 07, 2018 08:18
Hello,
Why you said it's booting problem??
It may be something else specially that error 4F2D does not relate to booting sequence.
Did you check the error log of DL?
I can help you to solve your issues
to rate and post a comment
(4)
Roberto Angel Vaca Valverde
re: re: Innova 2000 booting problem
November 08, 2018 09:55
Dear mina.
Yes on the dl logfile we can see errors with the detector communication
to rate and post a comment
(4)
Roberto Angel Vaca Valverde
re: re: Innova 2000 booting problem
November 08, 2018 09:55
Dear mina.
Yes on the dl logfile we can see errors with the detector communication
to rate and post a comment
(16)
GH
re: Innova 2000 booting problem
November 07, 2018 08:20
send me on g.masri@des-carte.com
I can help you
to rate and post a comment
(16)
GH
re: re: Innova 2000 booting problem
November 09, 2018 08:53
check the fiber cables from detector up to RTAC, and also check the C80 Mezzanine board
to rate and post a comment
(4)
Roberto Angel Vaca Valverde
re: Innova 2000 booting problem
November 12, 2018 03:30
Hi we found broken fiber insite de detector. Now no 4f2d or 4f2e, but our 4f48 boot error its on advantx console. On tbe service home page i cant see the positioner release, Unknow is showing
to rate and post a comment