GE LightSpeed Plus - STORELOG issue

May 28, 2015 01:31

Hi guys,

I've been trying to perform a STORELOG command on the system, and it sends me a PANIC message:

******
PANIC: CPU 1: KERNEL FAULT
PC: 0xa80000002021f130 ep:0xffffffffffffb898
EXC code:128, 'Software detected SEGV '
Bad addr: 0x0, cause: 0x8

Log inor Register

to rate and post a comment

(30)
(9)

Dean Kenney

re: GE LightSpeed Plus - STORELOG issue

May 30, 2015 08:07

There's more issues I'm sure but you hard drive has some issues

Log inor Register

to rate and post a comment

(136)
(35)

FranciscoG

re: re: GE LightSpeed Plus - STORELOG issue

May 31, 2015 09:48

Thanks Dean,
I even changed the Image Disk Drive, but the system didn't recognize it.

I'm still not sure what is the issue, just that I can not delete information from the Image Disk that is inside.

Log inor Register

to rate and post a comment

(12)
(3)

Luis Santos

re: re: re: GE LightSpeed Plus - STORELOG issue

June 01, 2015 08:48

If it did not recognize it, log in as a root, install 2nd disk then reboot login again as a root and do a image database reset.
It is recommended that you load from cold with new OS hard drive them install 2 disk.

Log inor Register

to rate and post a comment

(136)
(35)

FranciscoG

re: re: re: re: GE LightSpeed Plus - STORELOG issue

June 01, 2015 04:05

Hi Luis,

So everytime I change the 2nd hard drive (not the OS hard drive) is recommended to LFC, because I've logged in as root and restart the computer a few times. Thanks man!

Log inor Register

to rate and post a comment

(12)
(3)

Luis Santos

re: re: re: re: re: GE LightSpeed Plus - STORELOG issue

June 02, 2015 10:10

It is recomendé when you have the seen errors.
you can check the SYSLOG in /var/adm By typing more /var/adm/SYSLOG for errors on hard drives, the error will tell you Wichita hard drive

Log inor Register

to rate and post a comment