GE dash 3000
February 29, 2012 09:11
hi all.. we have a dash 3000 patient monitor with a problem that has no error code has been given initialization and it keeps on restarting...i think the software is corrupted...can you please help me...thanks..
to rate and post a comment
Paul Raffaelli
Re: GE dash 3000
February 29, 2012 05:43
If you would like to call our tech support, there is no charge whether you are an existing customer or not...call 305-470-8013 and ask for GE tech support...use my name if needed...Paul Raffaelli...no salesperson will get involved
to rate and post a comment
(1)
(5)
Thomas Wodarczak
Re: GE dash 3000
February 29, 2012 05:43
Have you tried a "Cold Start" on the unit? Do you have any thing displayed on the screen? It may be the Das Module. Try the Cold Start first.; Power off, press and hold record button, turn power on and hold record button until boot loader shows.
to rate and post a comment
(4)
Marcelo Tosi
re: GE dash 3000
June 16, 2014 05:24
Jonathan, I´m having same trouble.
Can you any solution?