Hello
I just encountered a problem when performing checkup procedure to a Somatom Scope Power from 2017. During the procedure is entering in a loop (repeating last scan until fatal error ). Any idea? Thank you in advance!
Please check the event log. You can get details of the error. If it is aborting scan in warm up cycle then it could be either x-ray tube or the x-ray generator issues. Is aborting in filament adaptation cycle or defective channel correction or in the air calibration cycle?
When you have got this error? When system start filament adaptation or deffective channels? If first - then problem can be with tube. If second- check detector or irs
SPECT, PET, CT and PET/CT equipment sales and service from OEM certified FSEs. Maintenance, repair, installation, relocation, and parts available with 24/7 support. Call 1-866-682-6331 or email sales@TTGimagingsolutions.com.
(1)
Marius Stoica
Checkup failure Siemens Somatom Scope 2017
November 16, 2021 01:56
Hello
I just encountered a problem when performing checkup procedure to a Somatom Scope Power from 2017. During the procedure is entering in a loop (repeating last scan until fatal error ). Any idea? Thank you in advance!
to rate and post a comment
(8)
(7)
JIFFY PAUL AKKARAKARAN
re: Checkup failure Siemens Somatom Scope 2017
November 17, 2021 10:11
Please check the event log. You can get details of the error. If it is aborting scan in warm up cycle then it could be either x-ray tube or the x-ray generator issues. Is aborting in filament adaptation cycle or defective channel correction or in the air calibration cycle?
to rate and post a comment
(67)
Gabriel Cool
re: Checkup failure Siemens Somatom Scope 2017
November 22, 2021 11:26
When you have got this error? When system start filament adaptation or deffective channels? If first - then problem can be with tube. If second- check detector or irs
to rate and post a comment