"Camera Iris Cal required" Displayed

Tags: C-Arm

(5)

Josh Thill

"Camera Iris Cal required" Displayed

February 27, 2015 04:42

I have a 9800 and 9900 OEC c-arm that are displaying this "camera Iris cal required" on the doghouse after having adjusted the KV on the camera. When looking in the service manuals it lists the "camera Iris cal required" but there is no information on how to calibrate this. Any help you be appreciated!

Thank you.

Log inor Register

to rate and post a comment

Lenny Place

re: Camera Iris Cal required Displayed

March 01, 2015 03:26

have you performed the Camera Iris cal using utility suite, if so if the values are the same for open and closed then you probably in the harness

1 person liked this
Log inor Register

to rate and post a comment

(10)
(3)

Natthakit Prongchai

re: Camera Iris Cal required Displayed

March 01, 2015 03:28

You need to use RUT or RUS application program (installed on laptop) to run this Camera Iris calibration.

1 person liked this
Log inor Register

to rate and post a comment

(5)
(1)

BRIAN TWOMEY

re: Camera Iris Cal required Displayed

March 01, 2015 03:31

After setting the iris for proper dose, you have to go into UtlitySuite, access RUT, goto to camera calibration and get iris stops. Don't loosen the stops as it says, just get iris stops, then write to system to update the new stop values. Don't forget to update data when you first access RUT so you have the correct data to write to system.

1 person liked this