Over 90 Total Lots Up For Auction at One Location - WA 04/08

AMX 4 no display

Tags: Radiology - R/F

(22)
(1)

Charley Russell

AMX 4 no display

July 21, 2021 09:30

I have an AMX 4 that I changed the battery charge board due to a fried resistor. I took the batteries out and charged each of the 9 batteries to 12.9-13 vdc. Reinstalled the batteries and the unit will not show anything on the display. The K 216 relay clicks but nothing else.

Log inor Register

to rate and post a comment

(1)

michael CARABOTTA

re: AMX 4 no display

July 22, 2021 10:52

Check and see if it makes an exposure. If it does replace the kv/ma display panel. (the small one that you increase/decrease kv/ma on). When this board goes bad it shorts out both panel displays. The long display board will still be ok.

Log inor Register

to rate and post a comment

(22)
(1)

Charley Russell

re: re: AMX 4 no display

July 23, 2021 11:52

Tried that, no rotor or xray. Tried two different controller boards, no go. Checked the power supply, and all of the outputs are good. Lights on controller board come on. The unit will not drive either.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs.

Medilab Global provides high-quality pre-owned imaging equipment, parts, and service to the healthcare industry worldwide. Our number 1 position is to provide excellent customer service.. 305-234-0084 | Email: info@medilabglobal.com | Hablamos Español

DOTmed Certified DOTmed 100 stats
advertisement

Upcoming Auction Events

NorthWest Supply
TIMED
Apr 08, 2024 (99 lots)