Our AGFA CP1000 auto processor is not drying films. Blower is working but infrared lamps don't seem to be. Do any of the fuses protect these. There is absolutely nothing in manual about lamps.
If you have the service manual the IR Lamps are in the schematic. I threw my manual away because that sucker is beyond end of life and we've converted our cutomers to AGFA CR or AGFA or Rayence DR.
The heaters are protected by a fuse. But you must check to see if there is voltage on the heaters. They don't run continuously, they only turn on when the film input roller senses a film being transported. Then it turns the pumps on to high speed and then and only then do the heaters turn on. Otherwise they are cold.
The heaters are controlled by a triac in the circuit.
But like I said earlier I no longer have a service manual and I doubt that any parts are available from AGFA for the CP1000.
SPECT, PET, CT and PET/CT equipment sales and service from OEM certified FSEs. Maintenance, repair, installation, relocation, and parts available with 24/7 support. 1-866-682-6331 or email sales@TTGimagingsolutions.com.
GERRY MCCULLOUGH
AGFA CP1000
March 04, 2020 09:54
Our AGFA CP1000 auto processor is not drying films. Blower is working but infrared lamps don't seem to be. Do any of the fuses protect these. There is absolutely nothing in manual about lamps.
to rate and post a comment
(36)
(1)
Ronald Swapp
re: AGFA CP1000
March 05, 2020 10:15
If you have the service manual the IR Lamps are in the schematic. I threw my manual away because that sucker is beyond end of life and we've converted our cutomers to AGFA CR or AGFA or Rayence DR.
The heaters are protected by a fuse. But you must check to see if there is voltage on the heaters. They don't run continuously, they only turn on when the film input roller senses a film being transported. Then it turns the pumps on to high speed and then and only then do the heaters turn on. Otherwise they are cold.
The heaters are controlled by a triac in the circuit.
But like I said earlier I no longer have a service manual and I doubt that any parts are available from AGFA for the CP1000.
to rate and post a comment
HEMSA
re: AGFA CP1000
March 23, 2020 10:32
Check that the thermal protector in the dryer rack is not open.
to rate and post a comment