Over 90 Total Lots Up For Auction at One Location - WA 04/08

Acuson antares 115 V supply change into 230 V.

Tags: Ultrasound, SIEMENS

(22)
(1)

Dr. Iqbal Malik

Acuson antares 115 V supply change into 230 V.

April 12, 2018 04:36

my acuson antares come with 115V and i want to convert it into 230V. Does this will effect on working of its boards or monitor. Plz. help in detail.

1 person liked this
Log inor Register

to rate and post a comment

(119)
(57)
(6)

Jim Carr

re: Acuson antares 115 V supply change into 230 V.

April 13, 2018 09:56

Changing the voltage on the AC Tray is easy and no other changes have to be made. The AC Tray is factory tested with 230VAC and are changed to 115VAC as needed.

a) Loosen one screw of the sheet-metal cover, located at the rear of the AC Power tray, to get access to the voltage switch.
b) Set the voltage switch position according to the illustration shown below the access hole.
c) Fasten the screw of the sheet-metal cover to hide the voltage switch.

Regards,
Jim Carr
AUE

2 people liked this
Log inor Register

to rate and post a comment

(22)
(1)

Dr. Iqbal Malik

re: re: Acuson antares 115 V supply change into 230 V.

April 15, 2018 11:08

Dear Jim sorry for late today I saw your reply and become very much glade to see your prompt reply. I am greatful.

Log inor Register

to rate and post a comment

(22)
(1)

Dr. Iqbal Malik

re: re: Acuson antares 115 V supply change into 230 V.

April 15, 2018 11:09

Hi Jim, thank you very much. Today i saw your prompt reply, sorry i was late bec. of my clinical practice. I will try to do myself and let you inform my success. Thanks again

Log inor Register

to rate and post a comment

(22)
(1)

Dr. Iqbal Malik

re: Acuson antares 115 V supply change into 230 V.

April 15, 2018 11:09

Hi Jim, sorry for late, today i saw your reply. Jim i am also greatful of your prompt reply to me as well as to other members too. Thanks

Log inor Register

to rate and post a comment

(22)
(1)

Dr. Iqbal Malik

re: Acuson antares 115 V supply change into 230 V.

July 27, 2018 09:43

Toshiba Xario XG power supply, can we change 120 V into 220 V

Log inor Register

to rate and post a comment

(119)
(57)
(6)

Jim Carr

re: re: Acuson antares 115 V supply change into 230 V.

July 27, 2018 04:59

Yes, it is also simple to change the input voltage settings for this system. There are 2 switches on the Transformer Tray, which is where the AC line cord is attached at the bottom rear of the system. With the power cord disconnected from the wall outlet, move each switch to the 220V - 240V position.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Repairs for less with FREE Transducer EVALS! CALL NOW 1-888-838-7488

If you have a broken or defective probe, why not attempt to repair before purchasing? This could save you $1,000’s! With a free evaluation we can let you know the cost of repair if applicable. Call today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

NorthWest Supply
TIMED
Apr 08, 2024 (98 lots)